Reader Mail
Wed. 12 Dec.07, 10:07
|
Different types of double inversion recovery (DIR, 2IR) sequences are used to improve the suppression of blood signal (black blood technique) or to null the signals from two different tissue types (e.g. white matter and cerebrospinal fluid).
The black blood technique (used in cardiovascular MRI) works with two inversion pulses, where the first pulse is nonselective and the second pulse is slice-selective. TI is set to a value at which the signal of the recovering inverted blood is zero (http://www.mr-tip.com/serv1.php?type=db1&dbs=Double%20Inversion).
The second technique (also named gray matter only) is used in brain imaging to improve the detection of lesions, for example in the diagnostic of multiple sclerosis. Two 180° pulses with different TI are used to suppress two different types of tissue simultaneously.
Hope this helps
|