6-1.5T MAGNETS, DIFFERING GRADIENTS
What parameter, if any, can compensate between magnets of equal field strength but very different gradient strengths? A T2 FAT sequence on one magnet can look very different on another magnet using same parameters.
 
 

Elise Gough RT(R)(CT)(MR)

Steven Ford

Thu. 3 Mar.11,
20:28
I assume that you mean a t2 fat suppressed sequence. Differing gradient strengths have only an indirect effect on these images. The fat saturation sequences require additional pulses which take time to execute; stronger gradient systems can execute these pulses faster.

If you see different results, it can be caused by a number of factors; if you can describe the differences, that would be helpful. Generally speaking, the quality of the magnet homogeneity makes a big difference. If the small FOV scans (wrist) look different from magnet to magnet, that's probably not the cause.

You should ask your MRI applications specialist about this, and pay attention to the TE and bandwidth. Are the FOV and number of steps the same?

Elise Gough

Fri. 4 Mar.11,
13:56
Thank you for your response. Our group is strict about FOV, thickness/gap parameters being the same across all magnets. One magnet has slew rate of 77, FRFSE T2, TE 85,TR 3500 Classic Fat supressed images which are pristine, vs magnets of 120 slew rate same parameters look signal starved. So gradient strength makes a speed difference, not an image quality difference?
 
 

Elise Gough RT(R)(CT)(MR)

Renate Semrau

Mon. 7 Mar.11,
15:11
Slew rate, rise time and/or duty cycle are specific parameters to describe the performance of the gradient amplifier. Amplifiers with higher performance allow a faster slew rate (shorter rise time). Stronger gradients allow to reduce echo time, increase the bandwidth and/or use a smaller FOV by influencing the SNR. To use same parameters on magnets with different gradient performance may not be recommended, or you have to adjust all parameters to the level of your magnet with the poorest performance.

Steven Ford

Mon. 7 Mar.11,
15:24
If some of the system are signal starved, it's most likely not due to gradient field strength and speed, but it can be related to that. Like everything else in MRI, it's complex and interdependent.

You're best advised to talk with the apps specialist from the systems that are under-performing, and learn from them what the bandwidth is, and also check the shim. Is the fat suppression OK, but the signal weak, or is the suppression bad also? Do you get a graphic output of the linear (gradient)shim corrrection on the various systems?


It is well worth the money that they might charge, if it comes to that, in order to optimize the use of the systems and for your own education. If the apps person does not know why the other machine works so well, there should be someone higher up in the organization who can shed some light on your question.

Elise Gough

Mon. 7 Mar.11,
15:56
Thanks so much for the input. NOBODY in my group seems to have a handle on why images vary in quality from 8 different magnets. I'm being pressed to take position of imaging specialist and thru my own research we have gradient strengths from 50 slew rate up to 120. Radiologists complain neuro stinks on the 50 slew rate 1.5T magnet and it is the most pounded on magnet in our group doing hospital inpatients. So many variables including tech expertise.
 
 

Elise Gough RT(R)(CT)(MR)

Steven Ford

Mon. 7 Mar.11,
16:07
We maintain a lot of magnets. The leading cause of image quality problems is applications related. Nobody can possibly know all the nuances of pulse sequences by various vendors, software levels, etc. I've heard many times that a certain machine is no good, when in fact the sequences are inefficiently set up. Look there first.

Usually the techs welcome good training. If they don't want to be trained, then you have a problem, but it sounds like this issue is caused by other factors.

If there are differences in baseline quality between the machines, then compensate for that by other means. Signal starvation is easily remedied; keep the quality as consistent as you can and let time be the variable, if it comes down to that.





Elise Gough

Mon. 7 Mar.11,
16:54
Thank you for your logical advice. Just spoke with GE apps. "No amount of messing with MR parameters can compensate for weak gradients. Increased gradient strength=decreased echo spacing=less blurring". Coincidently, my DO just emailed me asking what the gradient strength is on the magnet the RADs are complaining about.Just MAYBE they'll upgrade it. I need to visit our sites.The only way I have to check other sites' parameters is importing exams from PACS to my modality console. Tedious, but telling. Again, thankyou for all the help.
 
 

Elise Gough RT(R)(CT)(MR)

Forum Overview

  Category:  Applications and Examinations
Reply to this thread
(login or register first)
What is - 
'Forum Notification Service'?  
Nothing is wrong with California that a rise in the ocean level wouldn't cure.
- Ross MacDonald
 
Share This Page
FacebookTwitterLinkedIn

MR-TIP    
Community   
User
Pass
Forgot your UserID/Password ?    



Look
      Ups




MR-TIP.com uses cookies! By browsing MR-TIP.com, you agree to our use of cookies.

Magnetic Resonance - Technology Information Portal
Member of SoftWays' Medical Imaging Group - MR-TIP • Radiology-TIP • Medical-Ultrasound-Imaging • 
Copyright © 2003 - 2024 SoftWays. All rights reserved. [ 28 March 2024]
Terms of Use | Privacy Policy | Advertising
 [last update: 2024-02-26 03:41:00]